Hi @AtMostafa ,

The Gen3 minicubes have neutral density filters in user accessible trays on the bottom of the cube which block 90% of the excitation light. For use with pyPhotometry you probably want to remove these filters, because pyPhotometry uses pulsed illumination which requires higher LED-on light intensities (though the average light intensity is low as the led is off most it the time).

To measure the light power at the fiber tip use the 2EX_2EM_continuous acquisition mode such that that LED is on continuously not pulsed.

You can use the current released version of the GUI with the gen3 minicubes but the actual led current will be 3x higher than shown in the GUI. The latest versio…